I have been charged with solving a odor problem in a laboratory. A variety of plumbing fixtures are upstream of an acid neutralizing tank. The Canadian Plumbing Code mandates that neutralizing tanks must discharge to the sanitary system through a trap:
"Where a fixture or equipment discharges corrosive or acid waste, it shall dischage into a neutralizing or dilution tank that is connected to the sanitary drainage system through: a) a trap or b) an indirect connection"
(National Plumbing Code of Canada (2006), 2.4.4.4)
The plumbing inspector demanded that none of the fixtures upstream of the neutralizing tank be trapped, as that would consitute a double trapped condition. Therefore, some laboratory sinks and some floor drains are untrapped. The problem arises when bulk lab waste is disposed through the laboratory sinks. Offensive odour wafts up through the adjacent floor drains, and generates complaints.
Has this situation been encountered before? I've heard of a rubber "backflow" type insert that can prevent odour, but that seems like a bit of a mickey mouse solution which treats the symptom and not the problem. I believe the problem to be improper venting of the neutralizing tank itself.
Any comments?
"Where a fixture or equipment discharges corrosive or acid waste, it shall dischage into a neutralizing or dilution tank that is connected to the sanitary drainage system through: a) a trap or b) an indirect connection"
(National Plumbing Code of Canada (2006), 2.4.4.4)
The plumbing inspector demanded that none of the fixtures upstream of the neutralizing tank be trapped, as that would consitute a double trapped condition. Therefore, some laboratory sinks and some floor drains are untrapped. The problem arises when bulk lab waste is disposed through the laboratory sinks. Offensive odour wafts up through the adjacent floor drains, and generates complaints.